AJR: Handheld devices show promise in emergency radiology
Rachel J. Toomey, BSc, a researcher at the University College Dublin School of Medicine and Medical Science in Dublin, and colleagues wrote that a total of 46 percent of attending physicians and trainees and 45 percent of radiologists use PDAs.
“Although the benefits of handheld devices in the daily routine of clinicians is not under debate, the accurate display of medical images is disputed and has not been extensively researched,” said Toomey.
Researchers compared the diagnostic efficacy of a PDA and iPod Touch against that of secondary-class monitors for each of two image types — wrist radiographs and images from CT of the brain. They gathered a total of 168 readings by examining radiologists of the American Board of Radiology.
"In the PDA brain CT study, the scores of PDA readings were significantly higher than those of monitor readings when all observers' readings are taken into account,” the authors reported. “No statistically significant differences between handheld device and monitor findings were found for the PDA wrist images or in the iPod Touch devices studies, although some comparisons did approach significance.”
"This study showed that important clinical information about a patient's condition can be made available to clinicians through display of radiologic images on handheld devices. This finding extends the potential of the devices beyond current applications such as teaching residents and organizing clinical commitments," said Toomey.
The authors acknowledged that a study with more cases is desirable to decrease the width of the study and allow more definitive conclusions, but because of practical constraints, recruitment of a large sample was not possible in the study. Also, the investigators compared only secondary displays and handheld devices, so "no conclusions can be made about the performance of handheld devices in comparison with primary workstations."
"The results suggest that the handheld devices investigated in this study may be comparable with secondary monitors for reporting findings on intracranial bleeds on CT images and fractured wrists on radiographs and may be of value in radiology, particularly for teleconsultation and emergency procedures," she concluded.